Camera study suitability

Discussion in 'IATF 16949:2016 - Automotive Quality Systems' started by russell nugent, Dec 19, 2022.

  1. russell nugent

    russell nugent Active Member

    Joined:
    Mar 1, 2022
    Messages:
    70
    Likes Received:
    12
    Trophy Points:
    7
    Have been aske recently to conduct an MSA study on a camera we use on a mould cell to check graphic position . The problem is it is a go no go camera and doesnt give values . Could use R&R but again very little variability as go / no go . Looking for recommendations on type of study to provide start up confidence on the camera .
     
  2. Bev D

    Bev D Moderator Staff Member

    Joined:
    Jul 30, 2015
    Messages:
    606
    Likes Received:
    664
    Trophy Points:
    92
    Location:
    Maine
    There are two things to consider: how to perform a simple attribute study and how to perform a more detailed study should the simple study fail.
    First you really should understand how the camera works (it isn’t a simple go-no go device. The ‘camera’ uses pixels of a particular size and compares the graphic position tot he target area and the objects outline…Or something similar…)

    A simle attribute study would be to challenge the ‘gray area’. Position the graphic close to the boundaries, both in and out. depending on how the part is positioned to the camera you should also challenge the positioning of the object. (Shadows, glare, optical viewing angles will matter.). Don’t worry about the math at this point. The point of thi is to understand the facade accept and false reject area. How ‘big’ is it? Then think about the results. Is the camera adequate for your needs?

    If the camera isn’t then you need to understand how it actually works and how it is failing to pick up misalignment or why it is rejecting correct positioning.
     
    tony s, Miner, yodon and 2 others like this.
  3. Miner

    Miner Moderator Staff Member

    Joined:
    Jul 30, 2015
    Messages:
    578
    Likes Received:
    493
    Trophy Points:
    62
    Location:
    Greater Milwaukee USA
    There are many so-called Go/No-go testers that actually collect variable data internal to the tester then display the results as Pass/Fail. In almost every case, I have been able to find someone that can access the variable data stored internal to the tester and perform a standard R&R study.
     
    tony s, Andy Nichols and Bev D like this.