Does Uncertainty influence the acceptance of the Instrument post-calibration
Discussion in 'Gage Calibration and Uncertainty' started by Ganesh Sundaresan, Nov 19, 2015.
  1. Ganesh Sundaresan
    66
    36
    Uncertainty, I understand, is the zone surrounding the Calibration result within which the true value is supposed to exist. This being the case, does an Uncertainty of +/- 3 microns makes an Instrument, that is Calibrated and accepted by a margin of just 2 microns, skeptical for further use?
    Or is my fundamental understanding wrong?
  2. Ganesh Sundaresan
    66
    36
    A quick bump.
    Thank you in advance.
  3. Qualmx
    454
    58
    Normaly I consider two possibilities.
    The first,i trust in my supplier,I give th em the process tolerance and the equipment,then after calibraton I re eive the uncertainity then I solve the well known uncertainty equation ,from there and according the gotten values I decide if instrument is still suitable for the measurements.
    The other option it is to let the supplier to decide if the instrument it is suitable or not,noy giving them the process tolerance,they take the tolerance from the manufacturer.
    In my case I choose the second option because my tolerances are not too gtight
    Ganesh Sundaresan likes this.
  4. drgnrider
    drgnrider Member
    Nov 30, 2015
    44
    Kansas, USA
    Generally if you use the 10:1 rule, the uncertainty, for the most part, becomes irrelevant. Otherwise, as Qualmx said, you have to do the math and decide if the uncertainty and margin will work for you.