EMC Test Plan IEC60601-1-2 4th Edition

Discussion in 'IEC 60601 - Medical Electrical Equipment Safety' started by nikolaos, Aug 11, 2015.

  1. nikolaos

    nikolaos Member

    Joined:
    Aug 11, 2015
    Messages:
    9
    Likes Received:
    4
    Trophy Points:
    2
    Location:
    Italy
    Hi All,
    I am preparing a EMC test plan for a ME according the 4th Edition of the IEC60601-1-2. Is there anyone here working on this to share the opinions?
    Thanks
     
  2. Achaos

    Achaos Member

    Joined:
    Aug 15, 2015
    Messages:
    6
    Likes Received:
    0
    Trophy Points:
    1
    Hi Nikolaos,

    what´s your problem? How can I help you? At the Moment I`m preparing a test plan for the 4th Edition.

    Achaos
     
  3. nikolaos

    nikolaos Member

    Joined:
    Aug 11, 2015
    Messages:
    9
    Likes Received:
    4
    Trophy Points:
    2
    Location:
    Italy
    Hi Achaos,
    I am using the EMC Test Plan (Annex G – 60601-1©IEC:2014).
    How to specify the Dwell time for each Immunity test?
    Thanks
    Nikolaos
     
  4. Achaos

    Achaos Member

    Joined:
    Aug 15, 2015
    Messages:
    6
    Likes Received:
    0
    Trophy Points:
    1
    Hi Nikolaos,

    it depends from the function to be tested and the monitoring device to detect the failure mode (e.g. display change rate). If you use a display to detect the failure mode the dwell time shall be larger than the display change rate (e.g. 1 sec).
     
  5. Pads38

    Pads38 Member

    Joined:
    Aug 6, 2015
    Messages:
    28
    Likes Received:
    22
    Trophy Points:
    2
    Location:
    UK
    I do not (yet) have Ed. 4 of the EMC standard but I think that the determination of dwell times, for the radiated and conducted immunity tests, remains unaltered.

    The rationale, for the 2007 version, at clause 6.2.3.2 d) and e) states:-
    So dwell times are dependant upon what your device is doing, and what functions are being monitored during immunity tests. (Effectively your "Essential Performance"). The drawback of lengthy dwell times, of course, is that it substantially lengthens the time taken for each frequency band.